(65 intermediate revisions by 6 users not shown) | |||
− | + | <metadesc>Biodiesel titration, Turmeric titration, Biodiesel titration instructions, Testing the pH level of vegetable oil, Indicator solution, Blank titration, Titration photos, Titration calculations, Titration results, titration colours </metadesc> | |
+ | [[File:Turmeric.jpg|250px|thumb|right|Various stages of the titration test.]] | ||
+ | [[File:25ml burette.jpg|250px|thumb|right|Laboratory standard titration equipment. By NigelB]] | ||
+ | If you make biodiesel from fresh, unused vegetable oil there is no need to titrate the oil. This is because unused oil contains little or no [[free fatty acid]]s, this allows base weights of (5 grams per litre of oil for [[NaOH]] and 7 grams per litre for [[KOH]]) to be used in the biodiesel reaction. | ||
− | + | Used vegetable oil will contain free fatty acids in varying amounts depending on how, and for how long, it has been used. | |
− | + | Free fatty acids play no part in the biodiesel reaction and need to be removed. This is done by adding additional catalyst with which it will react to form soap. The soap is later removed as part of the production process. | |
− | + | Titration is a test that determines the level of free fatty acids known as the acid value or acid number. | |
− | + | It is very important to obtain accurate titration results, as these will have a major bearing on the quality of the finished biodiesel. For best accuracy, two or more titration tests can be carried out and the results averaged, especially if the titration levels are high. Experience has shown that pH meters don't work with [[WVO]] and are no substitute for a simple titration using indicator solution. | |
− | + | __TOC__ | |
+ | [[File:Titration scales.jpg|250px|thumb|right|Digital titration scales with TARE function.]] | ||
− | The items required for the test are very basic, although you can go to the expense of laboratory equipment, most home brewers don’t bother. Keep an eye on the household rubbish, over time most bits and pieces required can be found and reused. As we are private individuals carrying out this process at home we are not obliged to label any storage bottles to the same degree as would be required in a commercial situation. It is however, good | + | The items required for the test are very basic, although you can go to the expense of laboratory equipment, most home brewers don’t bother. Keep an eye on the household rubbish, over time most bits and pieces required can be found and reused. As we are private individuals carrying out this process at home we are not obliged to label any storage bottles to the same degree as would be required in a commercial situation. It is however, good practice and very strongly recommended, to clearly and indelibly mark each storage container with its contents and for solutions you make up, add the date so you have an idea of when they may need replacing. |
− | The scales will need to have a minimum resolution of 0.1 grams. You will only need this degree of accuracy infrequently for making your titration reference solution. However, if the range of the scales is large enough, say up to 1Kg, they can also be used for weighing the | + | The scales will need to have a minimum resolution of 0.1 grams. You will only need this degree of accuracy infrequently for making your titration reference solution. However, if the range of the scales is large enough, say up to 1Kg, they can also be used for weighing the catalyst during biodiesel production which saves buying two sets of scales, although with large processors and high titrating oil it may be necessary to weigh the catalyst in two hits. |
− | The platen on this type of scale is small but it’s capacity can be increased by using a suitable jug to contain the chemical and then using the “tare” facility | + | The platen on this type of scale is small but it’s capacity can be increased by using a suitable jug to contain the chemical and then using the “tare” facility to reset zero. |
+ | [[File:Measuring equipment.jpg|250px|thumb|right|Selection of suitable measuring equipment.]] | ||
Beakers in the laboratory, but jars in the bio lab. You will require a minimum of two jars. One to hold the oil sample, a jam jar is perfect for this and a smaller jar for the titration reaction. A baby food or fish paste jar is more suitable for this operation. Although two is a minimum, it’s always handy to have several jars to hand for spent solutions and used syringes and pipettes etc. | Beakers in the laboratory, but jars in the bio lab. You will require a minimum of two jars. One to hold the oil sample, a jam jar is perfect for this and a smaller jar for the titration reaction. A baby food or fish paste jar is more suitable for this operation. Although two is a minimum, it’s always handy to have several jars to hand for spent solutions and used syringes and pipettes etc. | ||
=== Measuring cylinders === | === Measuring cylinders === | ||
As with beakers laboratory type measuring cylinders can be used but plastic measuring cups form the kitchen work equally as well. For titrations only small volumes of 10ml are required, so a medicine measure might be employed or a syringe, see below. For making reference solutions you will need to measure 1 litre accurately and this will be done more conveniently with a larger vessel. | As with beakers laboratory type measuring cylinders can be used but plastic measuring cups form the kitchen work equally as well. For titrations only small volumes of 10ml are required, so a medicine measure might be employed or a syringe, see below. For making reference solutions you will need to measure 1 litre accurately and this will be done more conveniently with a larger vessel. | ||
=== Syringes and pipettes === | === Syringes and pipettes === | ||
− | [[File: | + | [[File:Storage bottles.jpg|250px|thumb|right|Dropper bottle, IPA dispensing bottle and old gear oil bottle for ref. solution.]] |
− | It’s largely down to preference as to which you use, but you will need | + | It’s largely down to preference as to which you use, but you will need two, possibly three sizes. A 20ml syringe will be useful if you don’t use the 10ml measuring cup mentioned above and something like 3ml plastic pipettes for measuring the oil sample and reference solution. If you have high titrating oil, a syringe of around 10ml may be more convenient. |
=== Dropper bottle === | === Dropper bottle === | ||
A small dropper bottle will be needed to store and dispense pH indicator solution. | A small dropper bottle will be needed to store and dispense pH indicator solution. | ||
Line 50: | Line 50: | ||
Line 62: | Line 61: | ||
+ | [[File:Tumeric box and solution.jpg|250px|thumb|right|Turmeric and solution in dropper bottle.]] | ||
+ | [[File:Weighing 1 gram of NaOH.jpg|250px|thumb|right|Weighing NaOH for ref. solution.]] | ||
− | + | See also: [[Titrating with turmeric]] | |
− | + | Indicators are substances that change colour at a certain pH. All indicator solutions are made by adding a little bit of an indicator to a lot of alcohol. They change colour when the solution they are in changes to a specific pH level. The two most common indicators used for making biodiesel are [http://en.wikipedia.org/wiki/Phenolphthalein phenolphthalein], and [[Titrating with Turmeric|turmeric]]. | |
+ | Phenolphthalein is the industry standard for titrating fats and oils to determine the percentage of free fatty acids they contain. It must be handled with care because it is an irritant. Phenolphthalein changes from colourless to pink at a pH 8.2 | ||
+ | Turmeric is by no means new as an indicator and references to it can be found in chemistry texts. Because turmeric is readily available in the spice section of most supermarkets, is cheap and lacks the irritant effect of phenolphthalein, its use as an indicator is preferred by the majority of home biodiesel makers. Turmeric changes colour from yellow to orange/red at a pH 8.5, close to that of phenolphthalein. | ||
+ | As turmeric is safe and readily available, this page will concentrate on its use. It needs to be made into a solution as follows: | ||
+ | In a clean jar mix isopropyl alcohol and powdered turmeric in the ratio of 5 parts isopropyl alcohol to 1 part turmeric by volume and mix well. The proportions are not too critical, but you should end up with a bright yellow solution with the turmeric powder settled at the bottom. Decant off the liquid into your dropper bottle and dispose of the dregs. For each titration you will only use a few drops so you don’t need to make up a massive quantity. As the rate of use is slow it is advisable to make fresh solution every few months to ensure that you always obtain accurate results. | ||
− | + | This needs to be a very accurately made solution of your catalyst, be it NaOH or KOH and distilled or deionised water. | |
− | This needs to be a very accurately made solution of your | ||
− | Weigh out 1 gram of | + | Weigh out 1 gram of catalyst to an accuracy of 0.1gram (this is why you require scales of a similar accuracy). It is best practice to use catalyst from the same batch as you will use to make your biodiesel. Ideally the catalyst should be 99% pure. |
− | Add the | + | Add the catalyst to the water and stir carefully until completely dissolved. |
− | A more accurate method of mixing a reference solution is to use the procedure above but use 10 grams of | + | A more accurate method of mixing a reference solution is to use the procedure above but use 10 grams of catalyst to 1 litre of distilled water. Then take 100ml of this solution and mix with a further 900ml of distilled water. |
− | For each titration you will only require a very small quantity of reference solution, typically a maximum of 10ml, so a one litre batch will give you at least 100 tests depending on the titration level of your oil. As with the indicator solution it is best practice to make a fresh solution after a few months to ensure the accuracy of the test. Similarly, if you purchase a new supply of | + | For each titration you will only require a very small quantity of reference solution, typically a maximum of 10ml, so a one litre batch will give you at least 100 tests depending on the titration level of your oil. As with the indicator solution it is best practice to make a fresh solution after a few months to ensure the accuracy of the test. Similarly, if you purchase a new supply of catalyst for biodiesel production, then you should make a new reference solution using the new catalyst. |
+ | ===Preparation=== | ||
+ | [[File:Titration kit.jpg|250px|thumb|right|Titration kit laid out ready for use.]] | ||
+ | Safety first. To carry out titration tests you should wear rubber or latex gloves, long sleeves and suitable eye protection. Assemble all the equipment and chemicals on a clear, stable work surface and be clear in you mind as to how you are going to proceed. | ||
+ | Take a small sample of oil. The sample should obviously be from the batch of oil you are about to convert to biodiesel. As most people collect oil in small batches from different sources and combine them to make one batch, it is preferable to have previously loaded the oil into the processor and mixed it thoroughly whilst heating. This ensures two things, firstly that the oil is a homogeneous and representative sample of the whole batch and secondly that it is not too cold to carry out the titration. For accurate results titrations should be carried out at approximately 20°C, having the oil pre-warmed will assist. | ||
+ | ===Titration overview=== | ||
+ | The object of the test is to ascertain how much additional catalyst will be required to convert the free fatty acids to soap. This is done by dissolving oil in [[isopropyl Alcohol]], adding indicator solution and then slowly adding the reference solution until the pH is brought to 8.5, at which point the indicator will change colour. As we are looking for quite small changes in pH levels it is imperative that all equipment is thoroughly clean before proceeding so as not to influence the result. | ||
+ | ===Blank titration=== | ||
+ | For accurate results a blank titration should be carried out. This is a process to ensure that the isopropyl alcohol if slightly acidic, will be neutralised and therefore not effect the accuracy of the titration. | ||
+ | Measure out 10ml of isopropyl alcohol in a jar or beaker. Add a few drops of pH indicator solution and ensure it’s thoroughly mixed. Take 2 or 3 ml of reference solution in a pipette or syringe and slowly add it to the jar drop by drop whilst stirring continuously. When the isopropyl alcohol reaches a pH of 8.5 the indicator solution will change colour. Turmeric will change from bright yellow to a dark crimson and phenolphthalein will change from yellow to pink. | ||
+ | If the colour change remains for 10 to 15 seconds, you have added enough reference solution and the isopropyl alcohol is ready for use. | ||
+ | <gallery caption="Blank titration sequence" widths="200px" heights="200px" perrow="3"> | ||
+ | File:Blank titration 1.jpg|10 ml isopropyl alcohol with a few drops of turmeric indicator solution. | ||
+ | File:Blank titration 2.jpg|With a few drops of reference solution added, on the point of colour change. | ||
+ | File:Blank titration 3.jpg|With one more drop of reference solution, complete colour chance and ready for titration. | ||
+ | ===Titration=== | ||
+ | Draw 1ml of oil into a clean pipette or syringe, add to the isopropyl alcohol and mix thoroughly. Draw a known volume of reference solution into a clean pipette or syringe, let’s say 5 ml. As with the blank titration add this to the isopropyl alcohol/oil mix drop by drop, stirring continuously until the indicator changes colour. If the colour change is stable for 20-30 seconds or so the titration is complete. | ||
+ | Note how much reference solution has been used. So if you have 2ml left in the pipette you have used 3ml and the titration value of the oil is 3. | ||
+ | It is good practice to repeat the process to ensure accuracy. You should expect the second result to be within 0.2 – 0.3ml of the first. The two values can then be averaged and this result taken as the titration value. | ||
+ | If the results vary significantly there has most likely been contamination during the titration and the tests should be re-run with clean equipment. | ||
+ | <gallery caption="Titration sequence" widths="200px" heights="200px" perrow="3"> | ||
+ | File:Titration 1.jpg|Blank titrated isopropyl alcohol with 1ml of oil added note the cloudy appearance and the colour change back to yellow. | ||
+ | File:Titration 2.jpg|With a few drops of reference solution added, colour change has started. | ||
+ | File:Titration 3.jpg|Additional reference solution added and colour change is complete. | ||
+ | Knowing the titration value we can now calculate the total weight of catalyst required for the batch of biodiesel. As mentioned at the top of the page the weight of catalyst required to convert new oil is a known value, 5 grams per litre for NaOH and 7 grams per litre for KOH. If you are just starting out with biodiesel production it’s safest to stick to these weights, but by varying the production process and by experimentation many home brewers have found these figures can be considerably reduced. | ||
+ | If your catalyst has a low purity often the case with KOH, it’s good idea to make an adjustment to the base figure to compensate for this. | ||
+ | Therefore if your KOH is 95% pure, the base weight should be increased as follows: | ||
+ | 7÷0.95 = 7.4 g | ||
+ | Conveniently the titration value equates to the number of grams per litre required to neutralise the free fatty acids. | ||
+ | We’ll stick with the generally accepted base weight figures for these examples. | ||
+ | Three pieces of information are needed for the calculation: | ||
+ | * The volume of oil to be processed … say 100 litres. | ||
+ | * The type of catalyst being used … say NaOH giving a base weight of 5 grams per litre | ||
+ | * And the titration value … from the above example, 3 i.e. 3 grams | ||
+ | Therefore the total weight of catalyst required will be the base weight (5) plus the weight derived from the titration (3) times the number of litres of oil to be processed (100), so: | ||
+ | (5+3)×100 = '''800 grams.''' | ||
+ | Another example using KOH at 95% purity. | ||
+ | * The volume of oil to be processed … say 80 litres. | ||
+ | * The type of catalyst being used … say 95% KOH adjusted to compensate for purity, giving a base weight of 7.4 grams per litre | ||
+ | * And the titration value … say, 0.5 i.e. 0.5 grams per litre | ||
+ | Therefore the total weight of catalyst required will be the base weight (7 adjusted to compensate for purity) plus the weight derived from the titration (0.5) times the number of litres of oil to be processed (80), so: | ||
+ | ((7÷0.95)+0.5)×80 = '''632 grams.''' | ||
+ | <br> | ||
+ | <br> | ||
+ | <br> | ||
+ | <br> | ||
+ | <br> | ||
+ | <br> | ||
+ | <br> | ||
+ | <br> | ||
+ | ==See also== | ||
+ | [[User:Bio-rich-time-poor|Bio-rich-time-poor]] 18:44, 28 November 2010 (UTC) | ||
Equipment required
Scales
Beakers
Measuring cylinders
Dropper bottle
Chemicals required
Isopropyl alcohol
Reference solution
Preparation
Titration overview
Blank titration
- Blank titration sequence
- Titration sequence
(5+3)×100 = 800 grams.