How strange that I'm posting this with 2 other threads ongoing about dropout tubes. It is actually coincidence.
PD1 (Paul) on here, sent me a sample of his bio to test a few days ago and his dropout result prompted me to make this post.
So, down to the sample. Water result was 400ppm- better than most and well in spec. Soap 44ppm again well within spec of 65ppm.
Dropout 4.5mls very high, even for a car that will cope with a high % of unconverted. After checking his result Paul got a result of 2.5mls. So why the big difference.
Paul had used a pipette to add his sample to the dropout tube and noticed that the level read about 27mls. Now to explain, this was one of Neisels tubes which read double the level on the low end. This isn't a problem as the reading is simply halved. The 10mls dispensed by the pipette should really read 20 on the tube, not 27.
Paul, thinking the pipette was incorrect poured some sample out of the tube until the level read 20, then added his methanol. This then gave the (incorrect) lower reading of 2.5mls dropout, explaining the discrepancy.
I calibrated several types of dropout tubes a few years ago and at the lower levels ( 0.1 to 2.0mls) the Neisel type tubes were more accurate than all the others I tested. However they arn't so accurate at the 10ml level (probably like most other tubes) so best not to use the tube calibrations for adding the sample.
I always use a pipette to add the sample, regardless of what the reading is on the tube. Even a cheap class 2 pipette is accurate to +/- 0.1mls far more accurate than a syringe. I realised a few years ago that the 10ml syringe I used actually dispensed closer to 11mls.